Teledyne API T803 CO2/O2 Analyzer Operation Manual Data Acquisition System (DAS & APICOM
137
instantaneous data points, e.g., for the new slope and offset (concentration
response) values at the end of a calibration. Zero and slope values are valuable to
monitor response drift and to document when the instrument was calibrated.
WARNINGS: Some data may be useful when stored if one of several warning
messages appears such as CO2TMW, O2TMPW (CO
2
or O
2
sensor temperature
warning). This is helpful for trouble-shooting by monitoring when a particular
warning occurred.
To edit the list of data parameters associated with a specific data channel, follow
the instruction shown in Section 7.1.5 then press:
Note DAS Trigger Events are firmware specific; a list of trigger events can
be found in Appendix A-5 of this manual.
7.1.5.3. EDITING DAS PARAMETERS
Data parameters are types of data that may be measured and stored by the DAS.
For each Teledyne API analyzer model, the list of available data parameters is
different, fully defined and not customizable. Appendix A-5 lists firmware
specific data parameters for the T803. DAS parameters include things like CO
2