Teledyne API T803 CO2/O2 Analyzer Operation Manual Setup Menu
71
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).
5.4.4. SETUP RNGE DIL: USING THE OPTIONAL DILUTION RATIO
FEATURE
This feature is an optional software utility that is used to compensate for any
dilution of the sample gas that may occur before it enters the sample inlet.
Typically this occurs in continuous emission monitoring (CEM) applications
where the sampling method used to remove the gas from the stack, dilutes the
sample.
Using the dilution ratio option is a 3-step process:
1. Select the reporting range mode and set the reporting range upper limit (see
Section 5.4.3).
Ensure that the upp
er span limit entered for the reporting range is the
maximum expected concentration of the non-diluted gas.