Teledyne API T803 CO2/O2 Analyzer Operation Manual Setup Menu
97
To activate the TEST Channel and select a function (in this example SAMPLE
PRESSURE), press:
SAMPLE CO2 RNG=20.00 % CO2=XXX.XX
<TST TST> CAL SETUP
DIAG TEST CHAN:NONE
PREV NEXT ENTR EXIT
DIAG SIGNAL I/O
PREV NEXT ENTR EXIT
DIAG TEST CHAN OUTPUT
PREV NEXT ENTR EXIT
Continue pressing NEXT until ...
Toggle to choose a
TEST channel
parameter
DIAG TEST CHANNEL:SAMPLE PRESSURE
PREV NEXT ENTR EXIT
EXIT discards the new
setting
ENTR accepts the
new setting
SETUP X.X PRIMARY SETUP MENU
CFG ACAL DAS RNGE PASS CLK MORE EXIT
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SETUP X.X ENTER PASSWORD:818
8 1 8 ENTR EXIT
Toggle to enter
the correct
PASSWORD
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).
07276B DCN6418