Teledyne T803 Tablet Accessory User Manual


 
Getting Started Teledyne API T803 CO2/O2 Analyzer Operation Manual
32
3.3. CONNECTIONS AND SETUP
This section presents the electrical (Section 3.3.1) and pneumatic (Section 0)
connections for setup and preparing for instrument operation.
3.3.1. ELECTRICAL CONNECTIONS
CAUTION
ELECTRICAL SHOCK HAZARD
Never disconnect PCAs, wiring harnesses or electronic subassemblies while the
analyzer is under power.
Note To maintain compliance with EMC standards, it is required that the
cable length be no greater than 3 meters for all I/O connections,
which include Analog In, Analog Out, Status Out, Control In,
Ethernet/LAN, USB, RS-232, and RS-485.
This section presents the electrical connections for AC power and
communications.
3.3.1.1. CONNECTING POWER
Attach the power cord to the analyzer and plug it into a power outlet capable of
carrying at least 10 A current at your AC voltage and that it is equipped with a
functioning earth ground.
CAUTION
ELECTRICAL SHOCK HAZARD
High Voltages are present inside the analyzers case.
Power connection must have functioning ground connection.
Do not defeat the ground wire on power plug.
Turn off analyzer power before disconnecting or connecting electrical
subassemblies.
Do not operate with cover off.
07276B DCN6418