9.2. MANUAL CALIBRATION CHECKS AND CALIBRATION
OF THE T803 ANALYZER
IMPORTANT
IMPACT ON READINGS OR DATA
ZERO/SPAN CALIBRATION CHECKS VS. ZERO/SPAN CALIBRATION
Pressing the ENTR button during the following procedure resets the
stored values for OFFSET and SLOPE and alters the instrument’s
Calibration. This should ONLY BE DONE during an actual calibration
of the T803.
NEVER press the ENTR button if you are only checking calibration.
9.2.1. SETUP FOR CALIBRATION CHECKS AND CALIBRATION
Connect the Sources of Zero Air and Span Gas as shown below.
Figure 9-1: Pneumatic Connections Using Bottled Span Gas
07276B DCN6418