Teledyne T803 Tablet Accessory User Manual


 
Teledyne API T803 CO2/O2 Analyzer Operation Manual Getting Started
51
after the 60-minute warm up period is over, investigate their cause using the
troubleshooting guidelines in Section 11.
Table 3-9: Warning Messages
MESSAGE DEFINITION
ANALOG CAL WARNING
The instrument's A/D circuitry or one of its analog outputs is not calibrated.
BOX TEMP WARNING
The temperature inside the T803 chassis is outside the specified limits.
CANNOT DYN SPAN
2
Remote span calibration failed while the dynamic span feature was set to ON
CANNOT DYN ZERO
3
Remote zero calibration failed while the dynamic zero feature was set to ON
CONFIG INITIALIZED
Configuration was reset to factory defaults or was erased.
CO2 ALRM1 WARNING
4
CO
2
concentration alarm limit 1 exceeded
CO2 ALRM2 WARNING
4
CO
2
concentration alarm limit 2 exceeded
CO2 CELL TEMP WARN
CO
2
sensor cell temperature outside of warning limits specified by
CO2_CELL_SET variable.
DATA INITIALIZED
DAS data storage was erased.
O2 ALRM1 WARNING
4
O
2
concentration alarm limit 1 exceeded
O2 ALRM2 WARNING
4
O
2
concentration alarm limit 2 exceeded
O2 CELL TEMP WARN
O
2
sensor cell temperature outside of warning limits specified by
O2_CELL_SET variable.
REAR BOARD NOT DET
CPU unable to communicate with the motherboard.
RELAY BOARD WARN
CPU is unable to communicate with the relay PCA.
SAMPLE FLOW WARN
The flow rate of the sample gas is outside the specified limits.
SAMPLE PRESS WARN
Sample gas pressure outside of operational parameters.
SYSTEM RESET
1
The analyzer was rebooted or the CPU was reset.
1
Does not clear after power up.
2
Clears the next time successful zero calibration is performed.
3
Clears the next time successful span calibration is performed.
4
Only active if the Concentration Alarm Option is installed
07276B DCN6418