Teledyne T803 Tablet Accessory User Manual


 
Teledyne API T803 CO2/O2 Analyzer Operation Manual Troubleshooting and Service
197
1. For Pressure related problems:
Measure the voltage across C1 it should be 5 ± 0.25 VDC.
If not then the board is bad.
Measure the voltage across TP4 and TP1.
With the sample pump disabled it should be 4500 mV ±250 mV.
With the pump energized it should be approximately 200 mV less. If
not, then S1, the pressure transducer is bad, the board is bad, or
there is a pneumatic failure preventing the pressure transducer from
sensing the absorption cell pressure properly.
2. For flow related problems:
Measure the voltage across TP2 and TP1 it should be 10 ±0.25 VDC.
If not then the board is bad.
With flow stopped (sample inlet blocked) the voltage should be
approximately 1V.
If the voltage is incorrect, the flow sensor is bad, the board is bad or
there is a leak upstream of the sensor.
07276B DCN6418