Teledyne T803 Tablet Accessory User Manual


 
Teledyne API T803 CO2/O2 Analyzer Operation Manual Calibration Procedures
169
9.4.2. FLOW CALIBRATION
The flow calibration allows the user to adjust the values of the sample flow rates
as they are displayed on the front panel and reported through COM ports to match
the actual flow rate measured at the SAMPLE inlet. This does not change the
hardware measurement of the flow sensors, only the software-calculated values.
To carry out this adjustment, connect an external, sufficiently accurate flow meter
to the sample inlet. Once the flow meter is attached and is measuring actual gas
flow, press:
SETUP X.X PRIMARY SETUP MENU
CFG DAS ACAL RANG PASS CLK MORE EXIT
SETUP X.X SECONDARY SETUP MENU
COMM VARS DIAG EXIT
SETUP X.X ENTER PASSWORD:818
8 1 8 ENTR EXIT
DIAG SIGNAL I/O
PREV NEXT ENTR EXIT
Continue pressing NEXT until ...
DIAG FCAL ACTUAL FLOW: 120 CC/M
0120 ENTREXIT
Toggle these buttons to
match the actual flow as
measured by the external
flow meter
EXIT discards the new
setting
ENTR accepts the
new setting
DIAG FLOW CALIBRATION
PREV NEXT ENTR EXIT
SAMPLE CO2 RNG=20.00 % CO2=XXX.XX
<TST TST> CAL SETUP
(The ACAL submenu in the Primary Setup Menu is a special configuration; consult factory).
07276B DCN6418